Ricardo Rivera

Agfa cr25

January 03, 2018 09:11

I need the software of cr25 to isnstall a new pc. if some one have and can help me thank you.

Log inor Register

to rate and post a comment

(25)
(5)
(2)

Marwan Kumsie

re: Agfa cr25

January 04, 2018 09:38

Dear Ricardo,
I have this problem many times.
-The IP address is not reset.get into the service and type the same IP address and confirm it.
-You are off the LAN ,some switch or connection are bad.
Regards

Log inor Register

to rate and post a comment

(62)
(10)

Ricardo Rivera

re: re: Agfa cr25

January 04, 2018 12:39

HI MARWAN, THANK YOU FOR YOUR TIME, I NEED RESET THE IP IN THE CR SCANER? IT´S OK OR IN THE PC. IF IS IN THE SCANER I DONT HAVE THE PASWORD, CAN YOU HELP ME?