AGFA CR-25

Tags: PACS / DR / CR / Digital

(69)
(6)

puma

AGFA CR-25

June 18, 2013 11:40

I have a error on agfa cr-25
Error Self-test failed Service 20454

Anyone have experience this trouble?

Log inor Register

to rate and post a comment

(42)
(4)

salman khan

re: AGFA CR-25

June 18, 2013 02:07

Hello Puma

20454
E_SCD_SCB_N15V_NOK
Meaning : The voltage -15V is not ok.
Reason :
Cure : Check voltage indicator LEDs, SMB, cables, ...