AGFA DRYSTAR 300 error MDM sensor
March 20, 2017 05:32
Hi colleagues,
my customer is getting error:
LAST MDM-SENS NOT OK
Any help is appreciated. Service manual, discks
Thank you all in advance
to rate and post a comment
CRAIG WAMBURA
re: AGFA DRYSTAR 300 error MDM sensor
March 22, 2017 10:24
MDM stands for"MACRODENSITOMETER". It is built-in to allow for constant image quality concerning Dmin, Dmax, sensitometry and vertical banding.
There are no adjustment procedures for the MDM but I would recommend you perform a full calibration of the MDM.
I can help with service manual and service software.
craig@telemed-technologies.com
to rate and post a comment
Ricardo Rivera
re: AGFA DRYSTAR 300 error MDM sensor
March 24, 2017 10:36
hi, the MDM is the densitometer, you can calibrate it in the control panel to level 2 or 3 maybe it everything, if not check the halogen lamp of the mdm, maybe is bad or not clean, it is in the right side of the machine.