Agfa drystar 3000 error 0cbc

(26)
(3)

EDILBERTO JUAREZ

Agfa drystar 3000 error 0cbc

December 29, 2011 09:13

Any body have idea who is couse for this error in Drystar 3000 by
AGFA, error ocurred when the system is comming up after have been pushed to acquire an image few second afther display error 6c87e.

Log inor Register

to rate and post a comment

(67)
(3)

farhan zeb

Re: Agfa drystar 3000 error 0cbc

December 30, 2011 09:18

Error Code --> 6c87e
E_T14_TH_CONNECT_TEST,
Meaning : During start-up (or interactively invoked), a dedicated
resistor measurement (covering about 20 elements), has
returned an error. This means the thermal head is not
connected.
Severe damage is not unlikely at this point.
Reason : a. Thermal head wiring is not connected - more
specifically this has got to do with the
data-cable (COS->AIC->head).
b. Everything is connected, but the cable is
damaged.
Cure : a. Insert the cable into the connector provided.
Check both links AIC->COS and AIC->head.
b. Replace the (flat-)cable causing the difficulty.

Log inor Register

to rate and post a comment

(17)
(1)

Jürgen Technikteam

Re: Agfa drystar 3000 error 0cbc

January 16, 2012 09:58

use the last backup
greeting
JH Technikteam

Log inor Register

to rate and post a comment