Error Code --> 6c44e
E_T14_PMC_1_START,
Meaning : PMC1 I/O-board could not be started successfully.
This board is responsible for driving all kinds of
actuators aboard the Drystar printer.
Reason : Board does not respond correctly during its
initialization phase (at start-up).
Cure : a. Check cables and board. Use CAN-analyser to
verify the CAN-communication if judged
necessary.
b. Check power-supply side as well, check the
power-cables leaving for the PMC-board.
c. Check BER-board sitting on top of the CPU-board.
Error 6C44E:
Meaning: PMC1 I/O-board could not be started successfully. This board is responsible for driving all kinds of actuators aboard the Drystar printer.
Reason: Board does not respond correctly during its intialization phase (at start-up).
Cure: a) check cables and board. Use CAN-analyser to verify the CAN-communication if judged necessary.
b) Check power supply side as well. check the power cables leaving for the PMC-board.
c) Check BER board sitting on top of the CPU board.
GCI offers complete MRI & CT sales & service with ROI analysis, in-house financing, contractor drawings, accreditation, training, marketing, and OEM-level support beyond the sale.
Leo Samba
Agfa drystar 3000 ERROR 6C44E
April 04, 2013 02:10
I have ERROR 6C44E CALL SERVICE on agfa drystar 3000. Need help
to rate and post a comment
Konstantin Kuznetsov
re: Agfa drystar 3000 ERROR 6C44E
April 05, 2013 10:05
Error Code --> 6c44e
E_T14_PMC_1_START,
Meaning : PMC1 I/O-board could not be started successfully.
This board is responsible for driving all kinds of
actuators aboard the Drystar printer.
Reason : Board does not respond correctly during its
initialization phase (at start-up).
Cure : a. Check cables and board. Use CAN-analyser to
verify the CAN-communication if judged
necessary.
b. Check power-supply side as well, check the
power-cables leaving for the PMC-board.
c. Check BER-board sitting on top of the CPU-board.
to rate and post a comment
tim cowell
re: Agfa drystar 3000 ERROR 6C44E
April 05, 2013 10:10
Error 6C44E:
Meaning: PMC1 I/O-board could not be started successfully. This board is responsible for driving all kinds of actuators aboard the Drystar printer.
Reason: Board does not respond correctly during its intialization phase (at start-up).
Cure: a) check cables and board. Use CAN-analyser to verify the CAN-communication if judged necessary.
b) Check power supply side as well. check the power cables leaving for the PMC-board.
c) Check BER board sitting on top of the CPU board.
to rate and post a comment
Konstantin Kuznetsov
re: re: Agfa drystar 3000 ERROR 6C44E
April 05, 2013 01:42
Forgive for that I outstripped you, Tim. I give you the way.
Good luck...
to rate and post a comment