Agfa Drystar 3000 error 6c87E

(7)
(1)
(1)

Cristian Pinto

Agfa Drystar 3000 error 6c87E

May 03, 2010 03:38

Any body have idea who is couse for this error in Drystar 3000 by AGFA, error ocurred when the system is comming up afther the TH Calibrating is present in display error 6c87e.

Log inor Register

to rate and post a comment

(373)
(66)
(15)

yuan bo

Re: Agfa Drystar 3000 error 6c87E

May 03, 2010 09:24

i can help you,please contact obnauy@163.com

Log inor Register

to rate and post a comment

(10)
(1)

mingyou fu

Re: Agfa Drystar 3000 error 6c87E

May 05, 2010 09:48

HELLO,
Error Code --> 6c87e
E_T14_TH_CONNECT_TEST,
Meaning : During start-up (or interactively invoked), a dedicated
resistor measurement (covering about 20 elements), has
returned an error. This means the thermal head is not
connected.
Severe damage is not unlikely at this point.
Reason : a. Thermal head wiring is not connected - more
specifically this has got to do with the
data-cable (COS->AIC->head).
b. Everything is connected, but the cable is
damaged.
Cure : a. Insert the cable into the connector provided.
Check both links AIC->COS and AIC->head.
b. Replace the (flat-)cable causing the difficulty.
by trying to drive another actuator.