Thomas Onyema
AGFA DRYSTAR 3000 INPUT JAM
February 08, 2011 11:46
Our drystar 3000 keeps jamming films persistently.The films barely leaves the film holding unit.
Any help will be appreciated.
Tom Onyema
to rate and post a comment
yuan bo
Re: AGFA DRYSTAR 3000 INPUT JAM
February 08, 2011 07:08
i have the service manual and service software, install software of dry3000, if any one need help, please contact obnauy@163.com
to rate and post a comment
eduardo medina
Re: AGFA DRYSTAR 3000 INPUT JAM
February 09, 2011 02:05
Hello, you need to check the holding unit, there are gears that moves a shaft with roller, probably adjust. Asuume that dosent reach the printing thermal head?
Eduardo
to rate and post a comment
Dante Nuevo
Re: AGFA DRYSTAR 3000 INPUT JAM
February 13, 2011 05:22
I have had that problem many times, it is usually dust in the roller.
It has nothing to do with the films, the drystar monitors the time the film takes to leave the holding unit. If roller motor doesn't have the correct force or the system doesn't make enough pressure over the film it moves slower and will not complete the process on time, so the printer will display FILM JAM error.
Extract the transportation roller clean it carefully, then clean the optical detector, lubricate the whole system with WD40 and finally use alcohol to clean the plastic roller.
That's all.
Good luck.
Dante Nuevo.