Over 150 Missouri Auctions End Today - Bid Now

AGFA DRYSTAR 4500M Problem

Tags: Laser and Dry Cameras, AGFA

(16)
(2)

KRV-4D

AGFA DRYSTAR 4500M Problem

March 02, 2013 03:01

Problem with Agfa 4500M ( Mamo ).

After 2 months without work. During power on, it seems the hard drive doesn't charge the software or data is corrupt. After 5 minutes, the machine display shows " SYSTEM ERROR " any idea ??

Thanks !!

Log inor Register

to rate and post a comment

(7)
(2)

Lloyd Bennett

re: AGFA DRYSTAR 4500M Problem

March 02, 2013 09:19

It sounds like the software is corrupted, the machine will probably not let you into service mode to reload the software from the hard drive. That means you will have to do an emergency software download from another computer. If that doesn't resolve the problem you have a bad hard drive or a board that's not letting the program finish loading.

Log inor Register

to rate and post a comment

(71)
(11)

bilal ahmad

re: re: AGFA DRYSTAR 4500M Problem

March 04, 2013 07:42

mail me its service manual and software
bilal.jaguar@gmail.com

Log inor Register

to rate and post a comment

(5)
(1)
(7)

Kiko Ramos

re: AGFA DRYSTAR 4500M Problem

March 06, 2013 08:19

Thanks Lloyd. The problem was fixed replacing the SCSI drive.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

OEM Philips & Siemens parts with Warranty!

Email P/N for pricing: sales@Relaxationinc.com (P) 813-925-1314, www.relaxationinc.com

stats
advertisement

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)