Over 500 Total Lots Up For Auction at One Location - NJ 06/15

Agfa DryStar 5300 Error system failure

Tags: PACS / DR / CR / Digital, AGFA

(11)
(6)

moncef dergali

Agfa DryStar 5300 Error system failure

June 01, 2020 11:02

dear all
i have a problem with the software of drystar 5300, error system failure"system self test failed; so please if you have it send to me

thanks a lot

Log inor Register

to rate and post a comment

(315)
(43)
(1)

Ronald Swapp

re: Agfa DryStar 5300 Error system failure

June 02, 2020 11:14

Your problem is not software related I keep telling you this.
You have a hardware problem, I recommend you contact either AGFA Support or your regional AGFA Service Representative.
You need to chase down the error number via the web gui, it will give you a clue as to what is going on.
The other is to monitor the boot process via a service laptop to determine what hardware component has failed.
Again, Software will not correct your problem.

Log inor Register

to rate and post a comment

(11)
(6)

moncef dergali

re: re: Agfa DryStar 5300 Error system failure

June 03, 2020 09:42

I asked for the software in relation to the technical manual which insists that this error is related to the software.
Thank you very much

Log inor Register

to rate and post a comment

(2)
(2)

JodecaRx

re: Agfa DryStar 5300 Error system failure

June 02, 2020 11:15

Hello, problem with software and or with the system disk.
Reinstall the new version of the software using the procedure in the service manual. Then place the backup of the equipment.
best regards

Log inor Register

to rate and post a comment

(315)
(43)
(1)

Ronald Swapp

re: re: Agfa DryStar 5300 Error system failure

June 03, 2020 09:40

There is no system disk!
This uses ram for the hd.

1 person liked this
Log inor Register

to rate and post a comment

(45)
(9)

CharlieWheel

re: Agfa DryStar 5300 Error system failure

June 02, 2020 11:16

Maybe your thindal/themis module it´s damaged. it is neccesary get access via Service port and check what occur when you turn on your printer.

Log inor Register

to rate and post a comment

(315)
(43)
(1)

Ronald Swapp

re: re: Agfa DryStar 5300 Error system failure

June 03, 2020 09:40

The post does not specifically point to an error message.
You cannot give a reasonable cure without the error number.

1 person liked this
Log inor Register

to rate and post a comment

(11)
(6)

moncef dergali

re: re: Agfa DryStar 5300 Error system failure

June 03, 2020 09:43

Thank you very much, I'll try