May 25, 2016 07:15

Having a print error. Printer starts up normally, prints 20% of the image then stops, gives an error reading "Film Jam". Film is easily cleared, then attempts to print again, which yields the same result.

If I print a sample or test image that is pre-loaded in the printer, 100% of the image prints and there are no film jams.

Any help would be appreciated!

Log inor Register

to rate and post a comment

(16)
(3)
(7)

Haider Ali

re: Agfa Drystar 5300 Print Error

May 25, 2016 05:25

reload the printer software. your problem will be solve.

Log inor Register

to rate and post a comment

(14)
(2)

David Guth

re: re: Agfa Drystar 5300 Print Error

May 26, 2016 09:39

Located the CF card, only to discover it does not have software on it. Should be version R1.9.2C1 Can anyone email me a copy of the software???
Thanks!

Log inor Register

to rate and post a comment

(332)
(69)

Gabriel Cool

re: re: re: Agfa Drystar 5300 Print Error

May 26, 2016 03:24

Write me your email i will send you software

Log inor Register

to rate and post a comment

(14)
(2)

David Guth

re: re: re: re: Agfa Drystar 5300 Print Error

May 27, 2016 04:56

I'd rather not post my email here, how about if you include the software as a link. Thanks

Log inor Register

to rate and post a comment

(102)
(39)
(1)

Konstantin Kuznetsov

re: Agfa Drystar 5300 Print Error

May 26, 2016 09:39

Try to send to the printer a task with smaller quantity the image. For example 1x1, 2x2, 2x3. Or send a task from other program or from other computer.

Perhaps you have incorrectly adjusted the DICOM printer driver on the computer?

Log inor Register

to rate and post a comment

(332)
(69)

Gabriel Cool

re: Agfa Drystar 5300 Print Error

May 26, 2016 09:41

This error has two solutions reload sw. if problem still present, change CF card and reload sw again

Log inor Register

to rate and post a comment

(14)
(2)

David Guth

re: re: Agfa Drystar 5300 Print Error

May 31, 2016 10:32

Thank you for all of your help! Had to load the software twice. It is now printing films perfectly!

Log inor Register

to rate and post a comment

(16)
(3)
(7)

Haider Ali

re: re: re: Agfa Drystar 5300 Print Error

May 31, 2016 12:39

Best Of Luck

Log inor Register

to rate and post a comment

(59)
(2)

Abu-Yusuf

re: Agfa Drystar 5300 Print Error

May 26, 2016 09:41

reload the software. there should be a sd card with software backup. Good luck

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your Trusted Source for Sony Medical Displays, Printers & More!

Ampronix, a Top Master Distributor for Sony Medical, provides Sales, Service & Exchanges for Sony Surgical Displays, Printers, & More. Rely on Us for Expert Support Tailored to Your Needs. Email info@ampronix.com or Call 949-273-8000 for Premier Pricing.

stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)