Error Code --> 6c841
E_T14_S8_MALFUNCTION,
Meaning : During a drum-calibration only a single
encoder-readout has been obtained. This is
not correct. A malfunction caused this
error-condition.
Reason : Possible malfunction or disconnection of
the drum-(slit-)sensor . As a result, since
no sensor-transitions are spotted, only a single
encoder read-out has been spotted.
Cure : a. Check sensor s8.
b. Replace the sensor.
c. Check the wirings.
d. Check the PMC-board functioning and release stamps.
Sometimes, we have the same error, but after off/on, it's ok.
From preventive maintenance through comprehensive cryogen services, BMI-PMI is your trusted partner for new/reconditioned sales, installation, and service of both facility-based and mobile MRI solutions. www.brownsmedicalimaging.com / 800-701-9729
Manuel Esquer
Agfa Drystar3000
October 07, 2013 12:44
I got error 6c841 on AGFA DRYSTAR3000. Any idea about solution?
to rate and post a comment
Irfan Malik
re: Agfa Drystar3000
October 07, 2013 02:09
6C841
Error Code --> 6c841
E_T14_S8_MALFUNCTION,
Meaning : During a drum-calibration only a single
encoder-readout has been obtained. This is
not correct. A malfunction caused this
error-condition.
Reason : Possible malfunction or disconnection of
the drum-(slit-)sensor . As a result, since
no sensor-transitions are spotted, only a single
encoder read-out has been spotted.
Cure : a. Check sensor s8.
b. Replace the sensor.
c. Check the wirings.
d. Check the PMC-board functioning and release stamps.
Sometimes, we have the same error, but after off/on, it's ok.
to rate and post a comment
Irfan Malik
re: Agfa Drystar3000
October 07, 2013 02:09
or Just replace the larger belt located on the left side of the printer
to rate and post a comment