The display is writing the error 958
than it says that a film has stuck in the machine.
I have opened all and i could find anything.
Please Help.
ardianbicaku@yahoo.com
Mr. Ardian, please check the photodiodes inside the printer, on the back of the chassis assembly. There are a few optical detectors that sometimes end dirty. Check a clean this, and if the problem persist, be free to contact me again.
Regards,
Gus.
Online auction of modern high-end laboratory and biopharma research equipment from the recent bankruptcy of Spexis AG in Allschwil near Basel, Switzerland. Start bidding today! closes on June 4. info@dome-auctions.com..+31(0)20 808 2880
Ardian Bicaku
AGFA SCOPIX LR5200 Error
May 11, 2007 11:38
The display is writing the error 958
than it says that a film has stuck in the machine.
I have opened all and i could find anything.
Please Help.
ardianbicaku@yahoo.com
to rate and post a comment
Gustavo Volpi
Re: AGFA SCOPIX LR5200 Error
May 15, 2007 08:26
Mr. Ardian, please check the photodiodes inside the printer, on the back of the chassis assembly. There are a few optical detectors that sometimes end dirty. Check a clean this, and if the problem persist, be free to contact me again.
Regards,
Gus.
to rate and post a comment
Ardian Bicaku
Re: AGFA SCOPIX LR5200 Error
May 17, 2007 01:49
thanks for advice, i tried that, but in the end it was a photo sensor that should be changed.
to rate and post a comment