I am needing Software and service manual for AGFA Solo I sent for International Children's Heart Foundation's Haiti project. I am getting a 203ce error at present. I hoping all I need is a cross over cable. For sustainability I really need the documentation. Thanks!

Roy Morris, CBET
Roy@babyheart.org

1 person liked this
Log inor Register

to rate and post a comment

(71)
(8)

TOM CHARLES

re: AGFA Solo CR

January 13, 2013 07:00

Hello Roy Morris

Your posiible causes may be

203CE
E_SCD_HT_READ_UNDER
Meaning : The measured high tension for the photomultiplier is more than
10% below the expected value.
Reason : - Cable from SCB board to high tension supply defective or
not connected properly.
- 24V supply or 24V cable to high tension supply defective or
not connected properly.
- High tension supply defective
- SCB board defective.
Cure : - Check cabelling and repair.
- Check / replace High tension supply (for photomultiplier).
- Check / replace SCB board.
- Check / replace 24V supply.