January 12, 2013 03:46
I am needing Software and service manual for AGFA Solo I sent for International Children's Heart Foundation's Haiti project. I am getting a 203ce error at present. I hoping all I need is a cross over cable. For sustainability I really need the documentation. Thanks!
Roy Morris, CBET
Roy@babyheart.org
to rate and post a comment
TOM CHARLES
re: AGFA Solo CR
January 13, 2013 07:00
Hello Roy Morris
Your posiible causes may be
203CE
E_SCD_HT_READ_UNDER
Meaning : The measured high tension for the photomultiplier is more than
10% below the expected value.
Reason : - Cable from SCB board to high tension supply defective or
not connected properly.
- 24V supply or 24V cable to high tension supply defective or
not connected properly.
- High tension supply defective
- SCB board defective.
Cure : - Check cabelling and repair.
- Check / replace High tension supply (for photomultiplier).
- Check / replace SCB board.
- Check / replace 24V supply.