AGM-philips need help

(3)

NGuyen Duy quy

AGM-philips need help

February 17, 2006 04:54

Hi all,

Please help to solve the error on AGM-philips (PN: M1024) with message" N2o, o2,n2o disturb".

Thank you in advance!
quentin

Log inor Register

to rate and post a comment

Muneer Abott

Re: AGM-philips need help

February 17, 2006 08:34

You have to go to the service menu, diagnostics screen, and find the problem exactly. It could be just defective selenoide, or a gas bench problem, but you must find out exactly which part is defetive from the service screen since Philips charges so much money for the parts. I have the service manual if you need it.

Log inor Register

to rate and post a comment

(3)

NGuyen Duy quy

Re: AGM-philips need help

February 20, 2006 12:53

HI all,

Who knows the supply source of IR head on AGM-philips PN: M1024. Also, I am looking for some more spare parts for Philips P.monitoring such as model V24, V24CT, C1, AGM ...