I have problems with display, monitor works with stripes, some black and some white. Aloka SSD 2000 has two ultrasound boards (EP 3445 E). I have changed places and is the same thing. Then ordered a new ultrasound board (EP 3312 M), the same problem again. When I remove the two boards monitor still works with stripes. When I disconnect the probe, lines disappear. The probe is ok. I'd appreciate an answer. Thanks
PS. If you give me an email address i can sent one photo of display
did you try to connect an external monitor and see is the monitor it self have a problem or not
and if you can send me images for the lines you mentiond about probe i may help you
elbaga85@hotmail.com
you mean sectors one black and one grey from the top to the bottom
may be a power supply problem if the two similar boards are good
try to remove the board on the left of these 2 boards and power on
if the power is good and no change due to this board then the problem in the sockets board or the mother board
Looking for the leader in urology? BK Medical is and we are the 3rd party experts. We sell BK and many other brands. Call today to discuss your urology equipment needs and save $1,000’s! Call Today: 1-888-838-7488 Email: sales@sentinelimaging.net
Gent Kosova
Aloka SSD-2000 Display Problem
January 30, 2014 02:29
I have problems with display, monitor works with stripes, some black and some white. Aloka SSD 2000 has two ultrasound boards (EP 3445 E). I have changed places and is the same thing. Then ordered a new ultrasound board (EP 3312 M), the same problem again. When I remove the two boards monitor still works with stripes. When I disconnect the probe, lines disappear. The probe is ok. I'd appreciate an answer. Thanks
PS. If you give me an email address i can sent one photo of display
to rate and post a comment
AHMEDELBAGA
re: Aloka SSD-2000 Display Problem
January 31, 2014 06:32
did you try to connect an external monitor and see is the monitor it self have a problem or not
and if you can send me images for the lines you mentiond about probe i may help you
elbaga85@hotmail.com
to rate and post a comment
Paulo Ferreira
re: Aloka SSD-2000 Display Problem
January 31, 2014 01:50
Please send photo to my e-mail paulo.ferreira@astrimed.pt
to rate and post a comment
emad aly
re: Aloka SSD-2000 Display Problem
January 31, 2014 03:19
you mean sectors one black and one grey from the top to the bottom
may be a power supply problem if the two similar boards are good
try to remove the board on the left of these 2 boards and power on
if the power is good and no change due to this board then the problem in the sockets board or the mother board
to rate and post a comment
Gent Kosova
re: re: Aloka SSD-2000 Display Problem
February 03, 2014 02:21
You solve my problem. Thanks
to rate and post a comment
Gent Kosova
re: Aloka SSD-2000 Display Problem
February 03, 2014 02:21
The problem was resolved. Was the problem of ultrasoun board Ep 3446 E. Thanks for your help