Tags: Radiology - R/F

Posted by an Anonymous User

AMX 4+ ERROR 435

June 26, 2019 11:10

Hallo. I need help. My GE AMX 4+ was working well until it started error 456. We managed to clear 456 and now we are getting error 435. We have checked the 6 conductors and the upper middle conductor was not pulling in. We tested its cable to another conductor and the cable from the middle conductor failed to pull in. We resolved the cable has an issue and we tested the cable which was discharging 110v current as required. We have checked the opt couplers and they all ok. What could be the problem with this system

Log inor Register

to rate and post a comment

(9)

Bill Diaz

re: AMX 4+ ERROR 435

July 01, 2019 10:22

Here is what I have for those errors. This will give you some idea of what needs to be checked. Bill

ERROR 466
1. During Calibration, Take Unit Out Of Calibration And Reset CPU Then
Restart Generator Calibration (MAS, KVP, Taps, And Filament Current Table)
2. Check Spark Gaps On HV Xformer For A 0.010” Setting
3. Check For Miswired Or Open Plug J2 From T2 On AMX1A4 Panel
To HV Xformer MA+ & MA-
4. Check For HV Xformer Spits
5. Check For XRT HV Spits
6. CPU Board May Be Defective
7. Insulate (On Older Units) Capacitor C5 From It’s Retaining Clamp (Ref. To S/N 3265)

ERROR 435
1. Check That All Contactors K1 Thru K6 On AMX1A4 Panel Pull In
When PREP Is Pressed
2. Check Installation Of Ribbon Cable From 1 KHZ Board To CPU Board
Or Opto-Couplers On 1 KHZ Board

Log inor Register

to rate and post a comment

(82)
(7)
(6)

Alfred M. Brown

re: AMX 4+ ERROR 435

July 01, 2019 10:21

ERROR 435
1. Check That All Contactors K1 Thru K6 On AMX1A4 Panel Pull In
When PREP Is Pressed
2. Check Installation Of Ribbon Cable From 1 KHZ Board To CPU Board
Or Opto-Couplers On 1 KHZ Board

Log inor Register

to rate and post a comment

(1)

(1)

Peter Richu

re: AMX 4+ ERROR 435

July 01, 2019 10:55

Hello. Check the cpu. Trace the connector supplying the tap signals from the inverter chip on the CPU board. There must be a broken connection from that chip to the 1KHZ connecor on the CPU. Or the flex cable.

Email at techs.mems@gmail.com for more help.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your one-stop source for all Pre-Owned Medical Imaging Equipment needs. πŸ“ž 305-234-0084

We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com

DOTmed Certified DOTmed 100 stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)