Hallo. I need help. My GE AMX 4+ was working well until it started error 456. We managed to clear 456 and now we are getting error 435. We have checked the 6 conductors and the upper middle conductor was not pulling in. We tested its cable to another conductor and the cable from the middle conductor failed to pull in. We resolved the cable has an issue and we tested the cable which was discharging 110v current as required. We have checked the opt couplers and they all ok. What could be the problem with this system
Check TAP1 FDBK through TAP6 FDBK signal of the
Generator and AEC Control port on sheet 3 location A 5 of
the CPU schematic.
Check connection to the 1kHz board.
Check tap select circuitry on the 1kHz board.
Check wiring to tap relay coils.
Error Code 35 Hex in the error list.
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
AMX 4+ ERROR 435
June 26, 2019 11:11
Hallo. I need help. My GE AMX 4+ was working well until it started error 456. We managed to clear 456 and now we are getting error 435. We have checked the 6 conductors and the upper middle conductor was not pulling in. We tested its cable to another conductor and the cable from the middle conductor failed to pull in. We resolved the cable has an issue and we tested the cable which was discharging 110v current as required. We have checked the opt couplers and they all ok. What could be the problem with this system
to rate and post a comment
Kirk Hannah
re: AMX 4+ ERROR 435
July 01, 2019 10:27
Check TAP1 FDBK through TAP6 FDBK signal of the
Generator and AEC Control port on sheet 3 location A 5 of
the CPU schematic.
Check connection to the 1kHz board.
Check tap select circuitry on the 1kHz board.
Check wiring to tap relay coils.
Error Code 35 Hex in the error list.