Over 150 Missouri Auctions End Today - Bid Now

AMX 4 Error code 45f

(1)

Tom Henderson

AMX 4 Error code 45f

June 01, 2006 02:06

Does anyone know what the error code 45f for an AMX is. Any help would be appreciated

Log inor Register

to rate and post a comment

(343)
(83)
(13)

Manuel Casanova

Re: AMX 4 Error code 45f

June 01, 2006 10:49

ERROR 45F
1. Check For Correct Wiring Between J5 Plug On FIL/KVP Board And HV Xformer
2. Check Spark Gaps On H/V Xformer (0.010)
3. Check That HV Cables Are Not Reversed
4. Check Wiring Of J4 Plug To FIL/KVP Board
5. Check That U18 On FIL/KVP Board Is #1489--NOT #14C89
6. Bad FIL/KVP BOARD (See Additional Notes #3)
7. FIL SHORTED (Check J2-25 On FIL/KVP Board)
8. Insufficient Filament Drive (XS-1 To XS-2 [HV Transformer] = Approximately 17 VAC @
52 KVP With PREP)
9. Loose J3 On Filter Board

Log inor Register

to rate and post a comment

(343)
(83)
(13)

Manuel Casanova

Re: AMX 4 Error code 45f

June 01, 2006 10:51

Open Filament In XRT (ERROR 45F) During Prep. Bad FIL/KVP Board (Q170, Q160
Or Q173 ShortedInverter Running At Maximum Output. CQA 964232.