I have an error 439 on my AMX 4, I cannot find a listing of error codes in our service menu. Does anyone know what this error is and how I could get a copy of the error codes. Thanks.
AMX 4 Error-439
Generator Fault
Probable Cause-The X-Ray On status signal was not asserted within 2ms after the Exposure start command signal was asserted.
Recommended Action-Check "EXP START CMND" signal of the Generator Control 1 port on sheet6,locationB7, and "X-RAY ON" signals of the Critical Status port on sheet3,locationD8 of the CPU schematic.
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
Tom Henderson
AMX 4 Error Codes
January 10, 2005 11:38
I have an error 439 on my AMX 4, I cannot find a listing of error codes in our service menu. Does anyone know what this error is and how I could get a copy of the error codes. Thanks.
to rate and post a comment
Tom Averbeck
Re: AMX 4 Error Codes
January 10, 2005 12:51
AMX 4 Error-439
Generator Fault
Probable Cause-The X-Ray On status signal was not asserted within 2ms after the Exposure start command signal was asserted.
Recommended Action-Check "EXP START CMND" signal of the Generator Control 1 port on sheet6,locationB7, and "X-RAY ON" signals of the Critical Status port on sheet3,locationD8 of the CPU schematic.
to rate and post a comment
Deon Armes
Re: AMX 4 Error Codes
January 17, 2005 12:54
There are 23 things to check in the error code listing for this error. email me at axsmedical@yahoo.com we can discuss them.
to rate and post a comment
Gordon Stokes
Re: AMX 4 Error Codes
March 10, 2005 11:29
Did you ever get a listing of all of the error codes for the GE AMX 4? I would be interested in such a list myself.
gordstok@sarmc.org
to rate and post a comment