AMX 4 Error Codes

Tom Henderson

AMX 4 Error Codes

January 10, 2005 11:38

I have an error 439 on my AMX 4, I cannot find a listing of error codes in our service menu. Does anyone know what this error is and how I could get a copy of the error codes. Thanks.

Log inor Register

to rate and post a comment

(1)

Tom Averbeck

Re: AMX 4 Error Codes

January 10, 2005 12:51

AMX 4 Error-439
Generator Fault
Probable Cause-The X-Ray On status signal was not asserted within 2ms after the Exposure start command signal was asserted.

Recommended Action-Check "EXP START CMND" signal of the Generator Control 1 port on sheet6,locationB7, and "X-RAY ON" signals of the Critical Status port on sheet3,locationD8 of the CPU schematic.

Log inor Register

to rate and post a comment

(110)
(10)
(3)

Deon Armes

Re: AMX 4 Error Codes

January 17, 2005 12:54

There are 23 things to check in the error code listing for this error. email me at axsmedical@yahoo.com we can discuss them.

Log inor Register

to rate and post a comment

(1)

Gordon Stokes

Re: AMX 4 Error Codes

March 10, 2005 11:29

Did you ever get a listing of all of the error codes for the GE AMX 4? I would be interested in such a list myself.
gordstok@sarmc.org

Log inor Register

to rate and post a comment

stats
DOTmed text ad

We provide sales, service & repair for portable ultrasound systems & probes

If you have a broken Sonosite, Mindray, GE, Philips, Hitachi, or Aloka system or probe contact us today. Free ultrasound & probe EVALS with 1 Year Warranty on all repairs. We repair all orders in 1-5 days. Call 844-644-5440 Email: Sales@Lenard.Healthcare

DOTmed Certified stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (537 lots)