amx 4 plus shows 03 04 failed when turned on,in service mode shows check sum error 8,passes calibration.controller pcb has been replaced.manufactured in 2006
Looking for the leader in urology? BK Medical is and we are the 3rd party experts. We sell BK and many other brands. Call today to discuss your urology equipment needs and save $1,000’s! Call Today: 1-888-838-7488 Email: sales@sentinelimaging.net
kevin mattingly
amx 4 plus errors
November 20, 2013 05:46
amx 4 plus shows 03 04 failed when turned on,in service mode shows check sum error 8,passes calibration.controller pcb has been replaced.manufactured in 2006
to rate and post a comment
Mohammed Abdullah
re: amx 4 plus errors
November 20, 2013 07:19
I have all manuals for this system contact me
magmesme@gmail.com
Regards
to rate and post a comment
Lenny Place
re: amx 4 plus errors
November 21, 2013 05:58
you need to perform the calibrations
to rate and post a comment
Steve Hilty
re: amx 4 plus errors
November 22, 2013 09:05
You need to do a complete calibration.