AMX4+, with new charger board, new CPU board and new batteries. Reloaded defaults and completed full calibration. When plugged in the unit will not charge and the unit will drive. Display reads x-rays disabled, drive only, charge recommended. With key switch off the unit still boots up, the only way to turn the unit off is with CB1. Relays on charger board activate when plugged in. We are getting a TEST 4 Failure on start up. Again, did complete full calibration after reloading defaults. Does anyone have any ideas?
Are the batteries fully charged? If not use the Checksum bypass when the unit is plugged in to charge them all the way then try again. I’ve had the same problem when my batteries were at 109ish volts.
Hope that works, if not let me know.
hi i noticed your posts and wonder if u can help me.
i replace charger board and batteries but when i try calibrating and reach to the part to REMOVE CHARGE CORD" and i do this the unit shuts off.. The unit does not turn n unless plugged into the wall outlet...any ideas??
We purchased a refurbished AMX4+ from a well known company as a backup unit.
The unit came fully charged and operational - sat idle for 2 months, when it came time to use there was no display or power. Battery voltage read aprox. 50.4vdc
Replaced batteries, calibrated (had to barnish TAPS relays to get through TAPS calibration - thought a refurbed unit would have been gone through more thoroughly?) Fully charged and calibrated, after a week or so techs went to use machine and no display or power.
Company is saying we damaged the batteries by leaving unit plugged in and charging...
DirectMed Imaging specializes in Equipment Sales, Parts, and Support. Parts are in stock and available for immediate delivery. In-room MRI Upgrades are available now. (855) 463-3727 or sales@directmedimaging.com
Mark Esquivel
AMX 4+ Problems
June 11, 2010 10:15
AMX4+, with new charger board, new CPU board and new batteries. Reloaded defaults and completed full calibration. When plugged in the unit will not charge and the unit will drive. Display reads x-rays disabled, drive only, charge recommended. With key switch off the unit still boots up, the only way to turn the unit off is with CB1. Relays on charger board activate when plugged in. We are getting a TEST 4 Failure on start up. Again, did complete full calibration after reloading defaults. Does anyone have any ideas?
to rate and post a comment
Joseph Hughes
Re: AMX 4+ Problems
June 14, 2010 04:28
Are the batteries fully charged? If not use the Checksum bypass when the unit is plugged in to charge them all the way then try again. I’ve had the same problem when my batteries were at 109ish volts.
Hope that works, if not let me know.
Regards,
Joseph
to rate and post a comment
Mark Esquivel
Re: AMX 4+ Problems
June 15, 2010 10:10
Recharged batteries, ran through the calibration again. Once I completed the last procedure (Field Light Timer) the unit started working.
Thanks for the help,
Mark
to rate and post a comment
F Khan
Re: AMX 4+ Problems
May 11, 2011 09:37
hi i noticed your posts and wonder if u can help me.
i replace charger board and batteries but when i try calibrating and reach to the part to REMOVE CHARGE CORD" and i do this the unit shuts off.. The unit does not turn n unless plugged into the wall outlet...any ideas??
to rate and post a comment
Nicholas Murphy
re: Re: AMX 4+ Problems
May 18, 2026 09:42
Can you explain the checksum bypass? I’m getting error where it it’s plugged in but stil says “recharge recommended “ as if it’s not charging.
to rate and post a comment
Dan Weger
re: AMX 4+ Problems
June 20, 2013 11:27
We purchased a refurbished AMX4+ from a well known company as a backup unit.
The unit came fully charged and operational - sat idle for 2 months, when it came time to use there was no display or power. Battery voltage read aprox. 50.4vdc
Replaced batteries, calibrated (had to barnish TAPS relays to get through TAPS calibration - thought a refurbed unit would have been gone through more thoroughly?) Fully charged and calibrated, after a week or so techs went to use machine and no display or power.
Company is saying we damaged the batteries by leaving unit plugged in and charging...