F Khan
AMX 4
May 11, 2011 09:37
My amx 4 wont turn on unless plugged in.. and when i reach to the REMOVE CHARGE CORD in the charger calibration the unit shuts off totaly when i unplug it from the wall.
Any ideas?
to rate and post a comment
Jose Morillo
Re: AMX 4
May 13, 2011 01:34
Check each battery voltage before and after charging.If voltage do not increase over the voltage of the battery, the charger is defective.
if baatery had been changed before fail appear, check polarity of each one, do not trust to the polarity color or label, sometimes mistakes had been maded during the manufacturing,
Check selector Charger, operation. Do you have voltage after switchs and reles conmute?
Check voltage in off and after turn it on,a suddenly drop mean a bad module
Best regards
to rate and post a comment
Arturo Papa
Re: AMX 4
May 19, 2011 11:08
Don't you have any error message?
to rate and post a comment
Youssef Desouky
re: AMX 4
October 22, 2013 08:58
If batteries measured voltage without plugging in the cord measured on the charger board TP9 and 10 is below 109 you are having a problem with either the charger board or the batteries need replacemnet