AMX4 Error 45F

(8)

(1)

C Walker

AMX4 Error 45F

October 17, 2008 10:47

Getting Error 45F after rotor start up checked connections on KVP/Filiment board and even changed out board still same error. Reading across XR Tube we get .3 oHms Common to Small and Large. I cecked the AC Voltage at XS-1 and XS-2 and we are getting 30 VAC and then it drops out. How can we verify tube is good.

Log inor Register

to rate and post a comment

(343)
(83)
(13)

Manuel Casanova

Re: AMX4 Error 45F

October 19, 2008 08:03

Walker:

ERROR 45F
1. Check For Correct Wiring Between J5 Plug On FIL/KVP Board And HV Xformer
2. Check Spark Gaps On H/V Xformer (0.010”)
3. Check That HV Cables Are Not Reversed
4. Check Wiring Of J4 Plug To FIL/KVP Board
5. Check That U18 On FIL/KVP Board Is #1489--NOT #14C89
6. Bad FIL/KVP BOARD (See Additional Notes #3)
7. FIL SHORTED (Check J2-25 On FIL/KVP Board)
8. Insufficient Filament Drive (XS-1 To XS-2 [HV Transformer] = Approximately 17 VAC @
52 KVP With PREP)
9. Loose J3 On Filter Board

If you need the service manual contact me asap.

Manuel