AMX4 KVP Cal Error1

James Mathis

AMX4 KVP Cal Error1

May 24, 2007 06:58

While performing KVP cal on an AMX4, The process continues up thru and including the 120KVP. The system does not ask for a value, says to remove the jumper, then displays "KVP Cal Error 1" after MAS UP is pressed. While vieing the waveform during cal, the initial wave form starts out low and ramps up to the required value at the 52, 64, and 85 settings. At the 120 setting, it starts off low, ramps up at about a 45 degree angle, but never actually reaches 120 before the system thinks cal is finished. The error then appears after jumper is romoved and MAS UP is pressed. The Ram chips and FIL Control board and HT cables have been replaced. Any ideas?

Log inor Register

to rate and post a comment

(52)
(5)
(8)

Samuel Sandlin

Re: AMX4 KVP Cal Error1

May 24, 2007 10:22


Check battery voltage, no more that 2 volts drop during prep cycle. If more than 2 volt drop the battery set is shot. After you get the system going no more than 20 volt drop at 100 KV,200 MAS.
Check the spark gap at the high voltage transformer with the lights off. Check for flashing light.
sam@amxrayservice.com

Log inor Register

to rate and post a comment

James Mathis

Re: AMX4 KVP Cal Error1

May 24, 2007 04:33

Problem was one bad battery in a set of new ones. Replaced bad battery and problem is resolved. Thanks for your response.

Log inor Register

to rate and post a comment

James Mathis

Re: AMX4 KVP Cal Error1

May 25, 2007 05:58

Problem was one bad battery in a set of new ones. Replaced bad
battery and problem is resolved. Thanks for your response.

Log inor Register

to rate and post a comment

(2)

tony villegas

Re: AMX4 KVP Cal Error1

May 24, 2007 10:38

Try to cal just using your KVP analyzer (insure it has been calibrated) - ditch the scope.

Log inor Register

to rate and post a comment

(2)

tony villegas

Re: AMX4 KVP Cal Error1

May 24, 2007 10:43

Sorry, I forgot to ask if you replaced the batteries. If not, than replace.

Log inor Register

to rate and post a comment

James Mathis

Re: AMX4 KVP Cal Error1

May 24, 2007 04:31

Yes, the batteries were new, but as it turned out, one of the batteries was defective. Problem solved. Thanks for your response

Log inor Register

to rate and post a comment

(244)
(19)

Staate Hayward

Re: AMX4 KVP Cal Error1

May 24, 2007 02:49

Double check the KVP tap relays. they have a small spring on a post that is held by a "c-clip" that may have broken. It will pull in but the contacts don't. Spring from a pen can be used and epxoy to hold.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

GE, Siemens, Philips, & Toshiba/Canon imaging parts, training & support!

DirectMed Imaging specializes in Equipment Sales, Parts, and Support. Parts are in stock and available for immediate delivery. In-room MRI Upgrades are available now. (855) 463-3727 or sales@directmedimaging.com

DOTmed Certified stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)