With just the information you giving is sounds like you've shorted you mAs board when you installed the meter. You can check the error log and is if there are any errors related to mA and kVp overload. Check the chathode high voltage cable for damage and the wells for arcing. Check the filaments and make sure they're not shorted to ground, or open. Beyond that I would replace the mAs board and start form there. Sense you're in the calibration mode the problem is going to be harder to fine.
One call to 440-263-6575 and your move will be covered.
We repair & store trailers. We build to your custom needs.
JDI Mobile Medical Manufacturing. Inc. We have your back.
Steve Oates
AMX4 Mas Calibration problem
January 12, 2010 12:49
AMX4 Mas Calibration. either breaker trips or no value in display
to rate and post a comment
Tom Walker
Re: AMX4 Mas Calibration problem
January 14, 2010 10:35
With just the information you giving is sounds like you've shorted you mAs board when you installed the meter. You can check the error log and is if there are any errors related to mA and kVp overload. Check the chathode high voltage cable for damage and the wells for arcing. Check the filaments and make sure they're not shorted to ground, or open. Beyond that I would replace the mAs board and start form there. Sense you're in the calibration mode the problem is going to be harder to fine.