January 10, 2013 10:04

what does 680 means on the display screen after refusing to turn ON?

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: amx4 mobile xray machine

January 10, 2013 05:29

I have never seen a error higher than the 400's is this a amx 4+?

Log inor Register

to rate and post a comment

(73)
(8)

Kim Schofield

re: re: amx4 mobile xray machine

January 11, 2013 08:26

The highest possible error code is 469

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: amx4 mobile xray machine

January 11, 2013 05:49

If the unit is giving a error code higher than the 400's than I would verify the power supplies on the CPU board check for ac noise also. The 680 is not within the normal rnage of error codes so it could be the CPU is giving a false code. If the supplies are good than I would replace the RAM chips and than perform a complete calibration. If you can not clear the 680 error replace the CPU

Log inor Register