MIT labs, experts in Multi-Vendor component level repair of: MRI Coils, RF amplifiers, Gradient Amplifiers Contrast Media Injectors. System repairs, sub-assembly repairs, component level repairs, refurbish/calibrate. info@mitlabsusa.com/+1 (305) 470-8013
dave wilson
amx4 model error code 321
July 25, 2011 03:02
amx4 model 2275938-9 error code 321. can anyone help?
to rate and post a comment
Juan de la Cruz Martinez
Re: amx4 model error code 321
July 25, 2011 05:16
Check Rotor Board
to rate and post a comment
Manuel Casanova
Re: amx4 model error code 321
July 26, 2011 02:59
DAVE:
ERROR 321
1. Items 1. And 3. From Error 320
2. Check For 27 VAC At J4-6 & 7 On Rotor Board
3. Check For 60 Hz Inverter Signal OK Signal At J1-5 On The Rotor Board
4. If Error 320 Occurs After Error 321, There May Be A Temperature
Problem On The 60 Hz Inverter
to rate and post a comment
Jaime Diaz
Re: amx4 model error code 321
August 20, 2011 10:40
The following are solutions for error 321 generator fault:
1. Check F93 On Rotor Board.
2. Check F7 on AMX1A3.
3. Check for 27VAC at J4-6 & 7 on Rotor Board.
4. Check for 60 hz Inverter Signal OK Signal At J1-5 on the Rotor Board.
5. If error 320 occurs after error 321, There may be a temperature problem on the 60 hz Inverter.
to rate and post a comment