Your Email:
Anode Overcurrent
September 27, 2013 10:09
On a GE CTLightspeed QXi got the following error during scan with large KV/MA or TIME HOST: OBC PROC: GENERATOR ERROR:183109 FUNCTION: XRAY GENERATION: KV LOOP KV FAULT DETECTED: ANODE OVERCURRENT FAULT LATCH ADDRESS: FFCFF3 HEX, BIT=D4 any idea about troubleshooting ? thank you
to rate and post a comment
Kim Schofield
re: Anode Overcurrent
September 28, 2013 09:19
Most likely the Anode tank. Check and clean the tube wells and try again.
Probably the anode tank. Clean and re-oil the cables at the tank and tube and try again.
Lightly used system available direct from the end user customer. MONARCH QUEST Featuring our most advanced, data-driven navigation software with AI-Powered algorithms & advanced imaging integration with the GE HealthCare OEC 3D C-arm.
Anode Overcurrent
September 27, 2013 10:09
On a GE CTLightspeed QXi got the following error during scan with large KV/MA or TIME
HOST: OBC PROC: GENERATOR ERROR:183109
FUNCTION: XRAY GENERATION: KV LOOP
KV FAULT DETECTED: ANODE OVERCURRENT
FAULT LATCH ADDRESS: FFCFF3 HEX, BIT=D4
any idea about troubleshooting ?
thank you
to rate and post a comment
Kim Schofield
re: Anode Overcurrent
September 28, 2013 09:19
Most likely the Anode tank. Check and clean the tube wells and try again.
to rate and post a comment
Kim Schofield
re: Anode Overcurrent
September 28, 2013 09:19
Probably the anode tank. Clean and re-oil the cables at the tank and tube and try again.
to rate and post a comment