APLIO 300

September 04, 2018 10:12

Hi, a need to help for aplio 300, code error TUS_IMG_33, somebody can help-me.
eciojuliano@gmail.com

thanks regards.

Ecio Juliano

Log inor Register

to rate and post a comment

(27)
(5)

Janicijevic Miroslav

re: APLIO 300

September 06, 2018 09:42

Hi Juliano,
Here is advice from service manual for this error massage.
US_IMG_33
Caution Configuration/periodic monitoring

Error in the imaging program. There may be a problem in the program. Restart the system.


I suggest you to create BackUp of the system (For safety) and after that perform reset the system to factory settings. Then perform "DB recreate".

I hope you can solve the problem whit these action.

GOODLUCK

Log inor Register