Aquilion 128 error

Tags: CT, TOSHIBA

(29)
(6)

MTC

Aquilion 128 error

April 09, 2015 10:09

Dear All,

We have error in Toshiba Aquilion 128 CT.

Error in Gantry Rotation

Contact your service representative.

code[0xb136000a] dev[0x00010001] Pno[8004]


Please advice us.

Log inor Register

to rate and post a comment

(29)
(6)

MTC

re: Aquilion 128 error

April 19, 2015 09:49

We found no voltage from TB2 output of CPU Box's power supply to Gantry. It caused Error in Gantry Rotation during tube Warm-Up. Also no voltage in couch.

Log inor Register

to rate and post a comment

(29)
(6)

MTC

re: re: Aquilion 128 error

April 26, 2015 11:55

"Error in Gantry Rotation" message is displayed during tube warm-up.

Also no power in the couch.

Please advice us

Log inor Register

to rate and post a comment