Yes, I have connected my laptop and did exposure test and in the Error description gives me those errors.
During normal operation, the warm up can not be done and always gives error "over ma"
Hi
It looks like tube problem
Doesn't mind how many slices it has
Depends how you use it , i got in my area same problem and some tube has also 500 000 slices or more
Concerning your problem , if I was to your place , i will check connection on generator and tube ,redo all connection with grease and oil/grease (depend on HV cable you use ) for the tube
If after this you get the same problem , i think you have to order some parts
best regards
anton
HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com
Aquilion 16 errors
September 02, 2012 08:02
In Aquilion 16 spellman generator, we have Errors HFG error & over ma(fixed), Can any body knows, what could be the problem?
Thanks for help
to rate and post a comment
Dominique Bultot
re: Aquilion 16 errors
September 03, 2012 11:09
Have you connected your laptop with the software XC test ?
to rate and post a comment
Nabil Naji
re: re: Aquilion 16 errors
September 03, 2012 07:40
Yes, I have connected my laptop and did exposure test and in the Error description gives me those errors.
During normal operation, the warm up can not be done and always gives error "over ma"
to rate and post a comment
Jose Soto
re: re: re: Aquilion 16 errors
September 03, 2012 08:52
check if some arc noise is present send error code to me ...kathzinstrumentos@hotmail.com
to rate and post a comment
Nabil Naji
re: re: re: re: Aquilion 16 errors
September 04, 2012 03:32
Hi Jose,
Please tell me how I can check if some arc noise is present?
The error code is : BFDFFFFFFF
to rate and post a comment
Yuriy Kuznetsov
re: re: re: re: re: Aquilion 16 errors
September 05, 2012 09:58
Check the X-ray
exposure waveform
using the Signal Show
function.
to rate and post a comment
Nabil Naji
re: re: re: re: re: re: Aquilion 16 errors
September 05, 2012 12:41
In the signal show,I can see that the ma signal is very high and more then 600ma and the kv signal is very low and around 1kv.
Any indication?
to rate and post a comment
QuantaMed
re: re: re: re: re: re: re: Aquilion 16 errors
May 24, 2015 10:04
Hi
This error is due a high current inside of the xray tube. So , defective part is the xray tube. Please, replaced It.
to rate and post a comment
Oleg Kuchin
re: Aquilion 16 errors
September 07, 2012 11:42
Its like Tube is broken. How long it working?
to rate and post a comment
Nabil Naji
re: re: Aquilion 16 errors
September 09, 2012 09:53
The tube worked two years with only 170,000 slices.
to rate and post a comment
juan pablo amado
re: re: re: Aquilion 16 errors
October 04, 2012 10:51
Dear Nabil, i`m having the same problem you have, could you tell me what was your conclusion? was the tube? i have the same slices number, 171 k
to rate and post a comment
Anton Spaci
re: Aquilion 16 errors
October 05, 2012 10:04
Hi
It looks like tube problem
Doesn't mind how many slices it has
Depends how you use it , i got in my area same problem and some tube has also 500 000 slices or more
Concerning your problem , if I was to your place , i will check connection on generator and tube ,redo all connection with grease and oil/grease (depend on HV cable you use ) for the tube
If after this you get the same problem , i think you have to order some parts
best regards
anton
to rate and post a comment
Pawel Kopysc
re: Aquilion 16 errors
May 06, 2015 03:08
Hi,
I have the same errors. Can You tell me what finally was broken?? Was it the tube?
Please email me kindly
kontakt@medikol.eu
Regards
Pawel
to rate and post a comment
ALSELWI
re: re: Aquilion 16 errors
May 19, 2015 01:49
It is the Tube.