Aquilion 64 error

Leonardo Sara da Silva

Aquilion 64 error

January 23, 2012 09:17

Hi,

i am from Brazil and have an Aquilion 64 scanner. After the warm up today, an error message appeared:

Error XC
code [0xb2640000] dev [0x00010002] Pno[1]

In the detais there are:

GTS detail error 4040404040....
DDmotor detail error 303030303....
Slide error detail 3030303030000000
Heigth error detail 3030303030303030...
Imudat error detail 30303030303030....
XC Status
mode (OR) , exec (UD), error (0x1a8
detail error 3b373f3f3f3f3f3f3....2f2f00

Can anyone help me with this?
thanks

Log inor Register

to rate and post a comment

(53)
(11)
(16)

Dominique Bultot

Re: Aquilion 64 error

January 23, 2012 10:03

Normally on Aq64, you have a small RS232 cable to connect on the back of gantry to have more detail on error, have you checked it ?
XC error can be any part from the XRay chain but also other trouble....better to check with XC status cable

dominique.bultot@skynet.be

1 person liked this
Log inor Register

to rate and post a comment

(78)
(13)
(1)

Ivano Murat

Re: Aquilion 64 error

January 23, 2012 05:37

Hi Leonardo!
Aquilion systems permit You to perform some tests only when using the long serial cable Mr. Bultot wrote about, while using a special made software to connect to the XC board inside the rotating part of the gantry, through some other boards.
So You'll need this cable, but if You cannot find it You can still use a normal serial null-modem cable (RX-TX twisted on the two sides of the cable), but be adviced this cable should be long enough to make it possible for You to perform the tests WITHOUT being exposed to the X-rays.
Furthermore, You'll need the software I wrote about, which is called XC-Test V.xx (depending on the Aquilion model You have).
So my advice is that You try and find the long serial cable as a first step, then we can discuss about what You might do with it.
In the meantime, have You tried to perform a seasoning, instead of a warm-up? Does it work correctly? Last question, has this system not been working for some days-weeks, before You did this last warm-up (maybe because of vacations or so)?
Best regards,
Ivano Murat

Log inor Register

to rate and post a comment

Ibrokhim Sultonzoda

Re: Aquilion 64 error

January 24, 2012 09:09

Dear Sir,

You should check in Generator board one fuse. Maybe this fuse is damage.

Best regards,
Ibrohim Sultonzoda

Ibrohim_007@yahoo.com

Log inor Register

to rate and post a comment

Robert Alcock

Re: Aquilion 64 error

January 24, 2012 02:57

Is this a Spellman or the older style SRU Toshiba generator? The error you listed is Abnormal input power / HFG error. (Spellman) Check the AC input voltage at TB1 of the AC chassis it should measure 200 VAC +/- 10% if good Measure the DC rail voltage at TB2 it should be 254 VDC or greater.

Log inor Register

to rate and post a comment

Leonardo Sara da Silva

Re: Aquilion 64 error

January 24, 2012 08:28

The engineer came and we think the problem is the inversor or the generator. We are going to change these thinks and see how it goes.

Log inor Register

to rate and post a comment

Ibrokhim Sultonzoda

Re: Aquilion 64 error

January 25, 2012 07:21

Dear sir,

XC - error mean, you don't have connection with generator. You should check generator board. In the generator board you must see one fuse, check this fuse.

Best regards,

Ibrohim S.

Log inor Register

to rate and post a comment

Ibrokhim Sultonzoda

Re: Aquilion 64 error

January 24, 2012 09:09

Dear Sir,

You should check in Generator board one fuse. Maybe this fuse is damage.

Best regards,
Ibrohim Sultonzoda

Ibrohim_007@yahoo.com

Log inor Register

to rate and post a comment

(7)

Andrey Khabylbkov

Re: Aquilion 64 error

February 09, 2012 08:16

Got a same problem, HV is Spellman 72 kW. Used program is Spag exe (XC test). In ERROR screen displays HFG error and AC fault. Yesterday I replaced AC unit, but nothing changed.
Can anyone help me with this?
thanks

Log inor Register

to rate and post a comment

Ibrokhim Sultonzoda

Re: Aquilion 64 error

February 10, 2012 09:14

Dear Sir,

You should check in Generator board one fuse. Maybe this fuse is
damage.

Best regards,
Ibrohim Sultonzoda

Ibrohim_007@yahoo.com

Log inor Register

to rate and post a comment

(7)

Andrey Khabylbkov

Re: Aquilion 64 error

February 11, 2012 11:46

Dear sir,
The fuses on every boards is OK. I checked them first. More idea?

Log inor Register

to rate and post a comment