Over 150 Missouri Auctions End Today - Bid Now

aquilion single slice

(1)

Daniel VIEYRA

aquilion single slice

October 24, 2012 02:09

I have error : XC DETAIL STATUS, DETAIL ERROR, 31363131313131313131212121212100 , IT OCURRS DURING PREPARE SCAN.

Log inor Register

to rate and post a comment

(53)
(11)
(16)

Dominique Bultot

re: aquilion single slice

October 24, 2012 06:13

Have you connected your laptop with service cable to the RTM ?

Log inor Register

to rate and post a comment

(88)
(36)

Anton Spaci

re: aquilion single slice

November 05, 2012 06:06

If i translate correctly , means
Door open
Input power abnormal
under kv fixed

does this message appears all the time ?
Do you have xc test ? for testing generator and tube
For the door open , you can already put a shunt between 5 and 6 on the TB where you connect warning lamps (back side of the gantry)

Log inor Register

to rate and post a comment

stats
DOTmed text ad

MRI & CT Support Worldwide. Buy Parts Fast with LASO.

Comprehensive medical imaging support. We supply essential parts & maintenance for most OEMs, buy/sell quality used equipment, and offer mobile rentals globally. Contact us: info@lasoimaging.com

stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)