I have error 2121-Xray exposure control on Canon Aquilion Vision. Fault indicated to the starter. I replaced the starter, error remained the same. Replaced the inverter with new XC board, still no luck.
Has any engineer managed to fix this?
HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com
Sithembiso Zondi
Aquilion Vision error2121
March 17, 2019 03:21
I have error 2121-Xray exposure control on Canon Aquilion Vision. Fault indicated to the starter. I replaced the starter, error remained the same. Replaced the inverter with new XC board, still no luck.
Has any engineer managed to fix this?
to rate and post a comment
Carlos Morales
re: Aquilion Vision error2121
March 19, 2019 10:20
If you changed those parts, tube may be bad. what moment error present? start up or tube ready?
to rate and post a comment
Sithembiso Zondi
re: re: Aquilion Vision error2121
March 20, 2019 09:54
After 10 second when SCON has finished loading aplication software.
We changed the tube in January this year because of bearing noise.
to rate and post a comment
Carlos Morales
re: re: re: Aquilion Vision error2121
March 21, 2019 09:45
I prefer you try whith original XC, may be configs are wrong.
to rate and post a comment
Sithembiso Zondi
re: re: re: re: Aquilion Vision error2121
March 22, 2019 10:35
I have done that. Error remains the same.
to rate and post a comment
Carlos Morales
re: re: re: re: re: Aquilion Vision error2121
March 26, 2019 10:13
does problem still? have you hearing anode to vibrate during power up?