Aquilion32 Help!

April 14, 2017 05:50

Hello!
Help solve the problem with the crescent-shaped artifact on Aquilion32

On the scan it is presented in the form of a white strip.
Appears after some time of work.
If you turn off the computer and turn it on again, the artifact disappears

Log inor Register

to rate and post a comment

(26)
(3)

QuantaMed

re: Aquilion32 Help!

April 17, 2017 04:43

Please did you check DAS power supply- Detector Temperature- Air calibration must not be old. Please can you test both reconstruction rack ? Also check if a red led lit on SCRT or GCIF PWB when artifact appears ( this system is MUDAT or SureCom?