Hello!
Help solve the problem with the crescent-shaped artifact on Aquilion32
On the scan it is presented in the form of a white strip.
Appears after some time of work.
If you turn off the computer and turn it on again, the artifact disappears
Please did you check DAS power supply- Detector Temperature- Air calibration must not be old. Please can you test both reconstruction rack ? Also check if a red led lit on SCRT or GCIF PWB when artifact appears ( this system is MUDAT or SureCom?
HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com
evg evgen
Aquilion32 Help!
April 14, 2017 05:50
Hello!
Help solve the problem with the crescent-shaped artifact on Aquilion32
On the scan it is presented in the form of a white strip.
Appears after some time of work.
If you turn off the computer and turn it on again, the artifact disappears
to rate and post a comment
QuantaMed
re: Aquilion32 Help!
April 17, 2017 04:43
Please did you check DAS power supply- Detector Temperature- Air calibration must not be old. Please can you test both reconstruction rack ? Also check if a red led lit on SCRT or GCIF PWB when artifact appears ( this system is MUDAT or SureCom?