Your Email:
sunil Kumar
artifacts philips ct
December 30, 2007 08:17
i am getting rings in air scans and steaks in patient scan.they are more pronounced in abdomen scans near pelvic region.head scans ,they are very light and almost non existent in upper cuts.can it be bad calibration??
to rate and post a comment
saju bhaskaran
Re: artifacts philips ct
December 30, 2007 01:42
Dear Sunil what system is this ?LX ,mx or brilliance? with regrads saju
Charles Psarreas
December 30, 2007 01:43
this sounds like a bad chan do a scout in the large field it you see a line artifact , this is your bad chan. what model is it?
Specializing in GE, Southeastern Biomedical delivers precise EKG monitor repairs and expert support you can trust. 888-310-7322.
sunil Kumar
artifacts philips ct
December 30, 2007 08:17
i am getting rings in air scans and steaks in patient scan.they are more pronounced in abdomen scans near pelvic region.head scans ,they are very light and almost non existent in upper cuts.can it be bad calibration??
to rate and post a comment
saju bhaskaran
Re: artifacts philips ct
December 30, 2007 01:42
Dear Sunil
what system is this ?LX ,mx or brilliance?
with regrads
saju
to rate and post a comment
Charles Psarreas
Re: artifacts philips ct
December 30, 2007 01:43
this sounds like a bad chan
do a scout in the large field
it you see a line artifact , this is your bad chan.
what model is it?
to rate and post a comment