I am getting error "error in optcont2(das count)" during scan and table movement is not possible.when we do reset the problem goesa nd scan is possible
In fact, OPCont is one of components of communication between static and rotates parts. Or your OPcont board is defective or one of signal is not correct. The best way is to connect the service cable on Main board to have precise error, see also in the service book of Asteion in the chapter MuDAT to have some sugegstions with LED
HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com
Punit Singh
Asteion multi error
September 15, 2012 12:29
I am getting error "error in optcont2(das count)" during scan and table movement is not possible.when we do reset the problem goesa nd scan is possible
to rate and post a comment
Yuriy Kuznetsov
re: Asteion multi error
September 15, 2012 01:52
write your e-mail
to rate and post a comment
Punit Singh
re: re: Asteion multi error
September 16, 2012 01:43
punit.globalmedical@gmail.com .the error comes after scano-table position not ok
to rate and post a comment
Dominique Bultot
re: Asteion multi error
September 16, 2012 07:58
In fact, OPCont is one of components of communication between static and rotates parts. Or your OPcont board is defective or one of signal is not correct. The best way is to connect the service cable on Main board to have precise error, see also in the service book of Asteion in the chapter MuDAT to have some sugegstions with LED