i am unable to scan and machine gives
error in recon process(No response to real time reconstruction setting).
the HW test is fail for link port single as well as multi "testing board test fail:load kernel to slot 12"
fibre channel and data collect exerciser test are passed.
Your problem is likely to be the mars pcb in slot 12. move the booards around and see if the errors moves to the slot where the faulty board has been relocated.
Ophthalmic-focused expertise in YAG systems including repair, maintenance and calibration handled with precision by trained professionals. Request service • sales@laserlocators.com • 1-877-924-2020
sunil Kumar
Asteion multi recon error
February 21, 2011 10:11
i am unable to scan and machine gives
error in recon process(No response to real time reconstruction setting).
the HW test is fail for link port single as well as multi "testing board test fail:load kernel to slot 12"
fibre channel and data collect exerciser test are passed.
please advise
to rate and post a comment
le hai nam
Re: Asteion multi recon error
February 22, 2011 08:36
Hello,
Did you check PIP PWB, FM PWB, Term PWB? You can test PIP PWB from CT manu.
to rate and post a comment
Jonathan Williams
Re: Asteion multi recon error
February 23, 2011 09:48
Your problem is likely to be the mars pcb in slot 12. move the booards around and see if the errors moves to the slot where the faulty board has been relocated.
to rate and post a comment