While taking table out in high speed it gets struck after moving some distance ,repressing out table switch(slow or fast) trips the table MCB.Slow movement out is ok with no problem .inward movement is OK.any suggestions??
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
Punit Singh
Asteion multi table problem
July 18, 2013 11:14
While taking table out in high speed it gets struck after moving some distance ,repressing out table switch(slow or fast) trips the table MCB.Slow movement out is ok with no problem .inward movement is OK.any suggestions??
to rate and post a comment
Anton Spaci
re: Asteion multi table problem
July 19, 2013 08:14
Do you have any message on screen
If not clean inside the couch
to rate and post a comment
Wang Zhixuan
re: Asteion multi table problem
July 20, 2013 08:23
You should first check the table,if it isn't problem you should test the table by the switches and bottoms in GTS
maintenance bosrd.
Mr.Wang zhixuan
362689810@qq.com
to rate and post a comment
Budi Gunawan
re: Asteion multi table problem
July 22, 2013 10:13
Replace GTSA pwb
to rate and post a comment