we have a problem in ATC generator from Gendix DEL . The 63A fuse is always blowing .SCR's and Tube and HV cables are ok .Any suggestions what to check
If you are sure the inverter is not shorted then check the main contactor I have had one contact go bad which caused another leg to draw to much current and blow the fuse
ATC 725 generator problem
May 13, 2017 09:08
we have a problem in ATC generator from Gendix DEL . The 63A fuse is always blowing .SCR's and Tube and HV cables are ok .Any suggestions what to check
to rate and post a comment
Lenny Place
re: ATC 725 generator problem
May 16, 2017 10:19
If you are sure the inverter is not shorted then check the main contactor I have had one contact go bad which caused another leg to draw to much current and blow the fuse