Bahram Hashemi

ATL 1000 Error 4

October 01, 2012 10:01

ATL 1000 ultrasound's power supply went out.
After replacing it, machine worked for 24 hours.
Then Error 4 appeared and it will not recognize the probes.

Machine was donated to Armenia, they do not have a service manual.

Any suggestion would be appreciated

Log inor Register

to rate and post a comment

Ricardo A. Sanchez

re: ATL 1000 Error 4

October 01, 2012 05:51

send me an email and I will send you the service manual

Log inor Register

to rate and post a comment

(42)
(8)

Bahram Hashemi

re: re: ATL 1000 Error 4

October 02, 2012 09:47

accuware1@yahoo.com

Log inor Register

to rate and post a comment

(100)
(45)

wang jia

re: ATL 1000 Error 4

October 01, 2012 07:34

hi;
I have the service manuals of HDI1000,HDI4000,HDI5000,UM9HDI,HD3,HD7,HD11,iu22,ie33 etc.E-mail:sdcwcywjj@163.com.

Log inor Register

to rate and post a comment

(100)
(45)

wang jia

re: ATL 1000 Error 4

October 01, 2012 07:34

send me an email and I will send you the service manual

Log inor Register

to rate and post a comment

(42)
(8)

Bahram Hashemi

re: re: ATL 1000 Error 4

October 02, 2012 09:47

accuware1@yahoo.com

Log inor Register

to rate and post a comment

(27)
(6)

Luis Vargas Carrillo

re: ATL 1000 Error 4

October 01, 2012 10:04

Check the voltages in the test points of the HV board (The expected voltages are labeled) you have to remove the probes port on the front and the board is in the left side

Log inor Register

to rate and post a comment

(42)
(8)

Bahram Hashemi

re: re: ATL 1000 Error 4

October 02, 2012 09:47

Thank you so much

Log inor Register

to rate and post a comment

stats
DOTmed text ad

We provide sales, service & repair for portable ultrasound systems & probes

If you have a broken Sonosite, Mindray, GE, Philips, Hitachi, or Aloka system or probe contact us today. Free ultrasound & probe EVALS with 1 Year Warranty on all repairs. We repair all orders in 1-5 days. Call 844-644-5440 Email: Sales@Lenard.Healthcare

DOTmed Certified stats
advertisement

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)