We provide sales, service, rental, and repair for ultrasound systems, probes, and C-arms. We have the largest inventory in the world. Contact us about our special offers on certified refurbished equipment! Call: 317-759-9210 Email: sales@probomedical.com.
muhammad shahid
ATL 5000 error # 002
September 15, 2014 09:16
machine reboot and given this error someone help of this error. thanx and best reqards
to rate and post a comment
Rey Gonzalez
re: ATL 5000 error # 002
September 15, 2014 06:19
removed the probes from the system, specially if you have a L12-5 and reboot the system, then plug the lineal probe in the center connector and try. for more help contact me rey@allultrasoundservices.com
to rate and post a comment
Konstantin Kuznetsov
re: ATL 5000 error # 002
September 15, 2014 06:22
"0002 FEC alert message -
The front end has stopped scanning because of some power monitor problem. This could be a hardware problem with the scanhead, the channel boards, the AIM, or the power supplies, but this also could be a system software problem. In reviewing the previous error messages, from the FEC, the board errorID should indicate if there is a specific problem with hardware."
Manuals to your scanner are available on the Internet ;-). Look for it as PHILIPS HDI-5000. Google to you in the help.