Greetings.
I have an atl hdi 5000 and I get the following error: diagnosis number 0002.
this error comes at a different time, at the beginning of the study, from the moment of choosing the trandcutor. please help.
note. How do I put the Spanish language in the atl
Error 002 ismost likely related to a problem with your aquisition section. SCANHEAD SELECTOR board could be defective or just dirty. Remove it, clean it with contact cleaner, re seat any CHANNEL BOARD, you see along with the AIM+ and try again.
Other problem could be the probes itselves. Try using the system with one probe at a time. If the problem appears with all the probes ypu may have a defective board.
Check the ERROR LOG of the unit to see any any problem highlighted by the system.
Error Code Descriptions on 17X.xx software:
******************
Error: 0002
Message: FEC alert message - The front end has stopped scanning because of some power monitor problem. This could be a hardware problem with the scanhead, the channel boards, the AIM, or the power supplies, but this also could be a system software problem. In reviewing the previous error messages, from the FEC, the board errorID should indicate if there is a specific problem with hardware.
******************
The most common problem is the power supply. I would check the modules producing 5 V - soldering, dirty contacts, loose terminals in modules ...
That error indicates the front end stopped scanning because of some power monitor problem. This can be caused by a scanhead with fluid infiltration, corrosion in the connector or an electronic part failing. It can also be caused by the channel boards, the AIM board, or the power supplies. It can also be a software problem. If it happens with different probes, I would suspect the power supplies or a channel board.
With the power off and the AC mains disconnected, try reseating the power supply and the front end boards. Then with the front cover off, monitor the power supply LEDs when the error happens.
Changing the language requires changing the Machine Options.
Hello Wellinton,
Error Code 0002:
FEC alert message - The front end has stopped scanning because of some power monitor problem.
This could be a hardware problem with the scanhead,
the channel boards,
the AIM, or the power supplies,
but this also could be a system software problem.
In reviewing the previous error messages, from the FEC, the board error ID should indicate if there is a specific problem with hardware.
Thank you very much everyone for your prompt response.
I only have one scan head so I can not determine if it is the probe.
but I'm going to dismantle the front-end feed and the cards, clean and comment.
note, I discard program problem?
I will do this and then comment if you have more options I thank you. If you can send me a service manual, it would be very helpful for me
wellintond@hotmail.com
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
wellinton de la cruz
atl hdi 5000 error
November 03, 2017 09:25
Greetings.
I have an atl hdi 5000 and I get the following error: diagnosis number 0002.
this error comes at a different time, at the beginning of the study, from the moment of choosing the trandcutor. please help.
note. How do I put the Spanish language in the atl
to rate and post a comment
Billy Sanchez Olivares
re: atl hdi 5000 error
November 04, 2017 04:23
Wellinton:
ATL HDI series doesnât support soanish laguage.
Error 002 ismost likely related to a problem with your aquisition section. SCANHEAD SELECTOR board could be defective or just dirty. Remove it, clean it with contact cleaner, re seat any CHANNEL BOARD, you see along with the AIM+ and try again.
Other problem could be the probes itselves. Try using the system with one probe at a time. If the problem appears with all the probes ypu may have a defective board.
Check the ERROR LOG of the unit to see any any problem highlighted by the system.
to rate and post a comment
Konstantin Kuznetsov
re: atl hdi 5000 error
November 04, 2017 04:24
Error Code Descriptions on 17X.xx software:
******************
Error: 0002
Message: FEC alert message - The front end has stopped scanning because of some power monitor problem. This could be a hardware problem with the scanhead, the channel boards, the AIM, or the power supplies, but this also could be a system software problem. In reviewing the previous error messages, from the FEC, the board errorID should indicate if there is a specific problem with hardware.
******************
The most common problem is the power supply. I would check the modules producing 5 V - soldering, dirty contacts, loose terminals in modules ...
Do you need a service manual?
to rate and post a comment
wellinton de la cruz
re: re: atl hdi 5000 error
November 05, 2017 04:16
please, if you can, send me the service manual to wellintond@hotmail.com
to rate and post a comment
Konstantin Kuznetsov
re: re: re: atl hdi 5000 error
November 06, 2017 09:57
Check your mail )
to rate and post a comment
Elio Tammarazio
re: re: re: re: atl hdi 5000 error
December 21, 2018 10:38
Please, could you send me the service manual cysller@gmail.com
to rate and post a comment
Hamza Omar
re: re: atl hdi 5000 error
July 11, 2018 10:26
hello
please can you send me the service manual and errors code .
to rate and post a comment
Hamza Omar
re: re: atl hdi 5000 error
July 11, 2018 10:26
hello
please can you send me service manual .
sahelalaaj@gmai.com
to rate and post a comment
Jim Carr
re: atl hdi 5000 error
November 04, 2017 04:28
That error indicates the front end stopped scanning because of some power monitor problem. This can be caused by a scanhead with fluid infiltration, corrosion in the connector or an electronic part failing. It can also be caused by the channel boards, the AIM board, or the power supplies. It can also be a software problem. If it happens with different probes, I would suspect the power supplies or a channel board.
With the power off and the AC mains disconnected, try reseating the power supply and the front end boards. Then with the front cover off, monitor the power supply LEDs when the error happens.
Changing the language requires changing the Machine Options.
to rate and post a comment
Marko Shevchuk
re: atl hdi 5000 error
November 05, 2017 03:57
Hello Wellinton,
Error Code 0002:
FEC alert message - The front end has stopped scanning because of some power monitor problem.
This could be a hardware problem with the scanhead,
the channel boards,
the AIM, or the power supplies,
but this also could be a system software problem.
In reviewing the previous error messages, from the FEC, the board error ID should indicate if there is a specific problem with hardware.
to rate and post a comment
wellinton de la cruz
re: atl hdi 5000 error
November 05, 2017 04:16
Thank you very much everyone for your prompt response.
I only have one scan head so I can not determine if it is the probe.
but I'm going to dismantle the front-end feed and the cards, clean and comment.
note, I discard program problem?