ATL HDI 5000 PROBLEM

Tags: Ultrasound

(70)
(3)

Garald jean louis

ATL HDI 5000 PROBLEM

October 05, 2021 10:10


I have an ATL HDI 5000 issue - need help - system suddenly stopped working - heard a snap inside just prior to failure . Have been receiving message system will shut down for high temperature lately . On the input board 5and 6 are lit - on PSM no over current light - the other one on . checked the voltage at the pins 0 - is that a PSM complete failure ? or is there a reset or something a fuse to change . help appreciated.
Thank you

Log inor Register

to rate and post a comment

(24)
(3)

heretohelp

re: ATL HDI 5000 PROBLEM

October 06, 2021 09:39

Since those two LEDs are lit on the input board, then the internal fuse of the input board is good and the 155 VDC is being sent to the card cage. The PSM has the LED lit, which indicates that the 155 VDC is present there. Since this is a good in/bad out situation, I would suspect that the power supply is bad. One other thing to check though is the input capacitor. On the back, to the right of the input board, you will find a large silver capacitor. Its job is to bump up the input AC. If it fails, there is not enough power to run the system. It rarely fails, but you should check it.

Log inor Register

to rate and post a comment

(70)
(3)

Garald jean louis

re: re: ATL HDI 5000 PROBLEM

October 07, 2021 10:42

Good I had that in thought but it seems to be so robust I was not convinced - will see that today .

Log inor Register

to rate and post a comment

(70)
(3)

Garald jean louis

re: re: ATL HDI 5000 PROBLEM

October 07, 2021 10:45

it is very annoying to remove that capacitor - it is easier too hook up another one out of the chassis to test . anyone has that capacitor specs please ?

Log inor Register

to rate and post a comment

(222)
(74)
(5)

Marko Shevchuk

re: ATL HDI 5000 PROBLEM

October 06, 2021 09:39

Hello Garald,
can you send me a pictures of message system
and from PSM of LED HVDC OK & OVP TRIG?

Log inor Register

to rate and post a comment

(222)
(74)
(5)

Marko Shevchuk

re: ATL HDI 5000 PROBLEM

October 06, 2021 09:39

Hi Garald,
All the voltages - + 15V - + 12V and other are formed by the corresponding modules.
The voltage HVDC = 175V is supplied to each module through fuses = F1 - F8.

Log inor Register

to rate and post a comment

(70)
(3)

Garald jean louis

re: re: ATL HDI 5000 PROBLEM

October 07, 2021 10:44

do you have a picture locating those fuses - or some indications

Log inor Register

to rate and post a comment

stats
DOTmed text ad

BK and other Urological Systems available. BK Probe repair available!!!

Looking for the leader in urology? BK Medical is and we are the 3rd party experts. We sell BK and many other brands. Call today to discuss your urology equipment needs and save $1,000’s! Call Today: 1-888-838-7488 Email: sales@sentinelimaging.net

DOTmed Certified stats
advertisement

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)