ATL HDI3000 error message

November 29, 2013 08:43

Hi,
We have a HDI3000 machine which has just started coming up with the error message:
SSPI MOP reported HW error
2303-23D0-0801-5B307832
SSP Hardware test failure

Can anyone tell me what this means? It started coming up a few times but would go away if the machine was switched off and on again but now it comes up every time.
Thank you!

Log inor Register

to rate and post a comment

Rick Reed

re: ATL HDI3000 error message

November 30, 2013 09:12

The SSP board (Sample space processor) is failing it's self-test.
Replace it.

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: ATL HDI3000 error message

November 30, 2013 09:12

Most likely the SSP (Single Space Processor) board is causingt the issue, I would first attempt to reseat the boards, check the SSP board ensuring all socketed chips are seated securely and reboot. If you still get the error I would replace the SSP

Log inor Register

to rate and post a comment

(29)
(6)

Jose Pacheco

re: ATL HDI3000 error message

December 01, 2013 07:50

There is a serial port on the rear of the system, plug in a serial cable and monitor it with a laptop using hyper terminal. The HDI 3000, 3500, and 5000 all push out serial data of the boot process and will give you access to the VX works software. Watch the boot up process and the failure will most likely show up in the serial output. 


Hyper terminal settings are below 


Serial Data 
Transfer 

• 9600 baud 

• 8 data bits 

• No parity 

• One stop-bit 

• Xon-Xoff protocol