Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

ATL HDI5000 PROBLEM

(34)
(8)

Ruben Cadima

ATL HDI5000 PROBLEM

March 03, 2012 12:06

Hi all
I have a problem with an ATL HDI5000 ultraosund scanner, the machine doesent power up and nothing is displayed in the monitor, there are LEDs on in olmost all PCBs but i have no idea which to review, also the coolers are activated and all the voltages seem to be normal. I guess the problem is in the host CPU but need help. Does any know how to troubleshoot this issue. Thanks beforehand

1 person liked this
Log inor Register

to rate and post a comment

(53)
(5)
(31)

Brian Seltzer

Re: ATL HDI5000 PROBLEM

March 03, 2012 03:53

Ruben,
your problem might be with the hard drive. Are you sure you checked all the power supply voltages? Do you hear High voltage come up on the monitor when you turn system on?

Log inor Register

to rate and post a comment

(74)
(15)

PIEBENG APPOLIN

Re: ATL HDI5000 PROBLEM

March 04, 2012 08:53

In the right side on machine you can found a MO disk with backup of system software
Put it in the drive an try boot your system with the MO inside
If your system boot with the MO so you have a HDD problem
Let the system boot normaly with MO
It can resolve your proble

Regards

Log inor Register

to rate and post a comment

(34)
(8)

Ruben Cadima

Re: ATL HDI5000 PROBLEM

March 05, 2012 05:53

Dear Brian
Thanks for your answer
Honesty i am not sure about the voltages but as a matter of fact, i dont listen the hig voltage activation to the monitor, i just reviewed the power suply voltages and thera re some not present but i donto know if they must at the moment when the machine doesent boot up

thanks beforehand
Ruben

1 person liked this
Log inor Register

to rate and post a comment

(74)
(15)

Sayat Karapetyan

Re: ATL HDI5000 PROBLEM

March 03, 2012 03:54

Hi Ruben,
I can help you.
Please write an e-mail.
With best wishes,
Sayat Karapetyan.
sayat1956@gmail.com

Log inor Register

to rate and post a comment

(72)
(5)

Muhammad yousaf

Re: ATL HDI5000 PROBLEM

March 04, 2012 08:56

Hi Mr.Ruben
you check the power section of your machine
first open the fron side of the machine and lock the power supply modular led status then i help u
Muhammad Yousaf Mirza

Log inor Register

to rate and post a comment

(2)

bensalah samir

re: ATL HDI5000 PROBLEM

October 02, 2017 04:35

Hello, I have a hdi5000 broken, the screen does not turn on but the keyboard and the led on the device are on, which I noticed the tht does not start.
if someone can help me solve this problem thank you
if someone has the service manual thank you to send it to me
samirmed007@yahoo.fr

Log inor Register