ATL ULTRAMARK 9 HDI CONTROLLER NOT RESPONDING

December 15, 2015 08:19

ATL ULTRAMARK 9 HDI DISPLAYS "CONTROLLER NOT RESPONDING"

ANY BOARD WITH AN IDEA

1 person liked this
Log inor Register

to rate and post a comment

fred74

re: ATL ULTRAMARK 9 HDI CONTROLLER NOT RESPONDING

December 15, 2015 05:08

This is often a Problem of the Hard Disk....in many case to can hear a mechanical "klick" from the HD 9,1 GB Fat

Log inor Register

to rate and post a comment

(121)
(33)
(9)

joe ssali

re: re: ATL ULTRAMARK 9 HDI CONTROLLER NOT RESPONDING

December 16, 2015 09:00

Thank you JANOSCH. TRUE It is the hard drive. When the CPU Starts booting, it stops on displaying 7 on both the upper and lower segments. IDEALLY IT IS SUPPOSED TO DOWNLOAD THE PROGRAM FROM THE HARD DISK WITHIN 27SEC AND THEN TURN TO 8 ON BOTH UPPER AND LOWER SEGMENTS.

Log inor Register

to rate and post a comment

(53)
(5)
(31)

Brian Seltzer

re: ATL ULTRAMARK 9 HDI CONTROLLER NOT RESPONDING

December 15, 2015 05:08

either cpu or HDD, we have both my friend

Log inor Register

to rate and post a comment

(121)
(33)
(9)

joe ssali

re: re: ATL ULTRAMARK 9 HDI CONTROLLER NOT RESPONDING

December 16, 2015 09:00

Dear Brian, thank you so much. TRUE, I also suspect the hard drive. The CPU looks fine because of the following boot squence F- 4- 7 BUT, 8 IS NOT DISPALAYED, MAY BE BECAUSE THE HARDDRIVE HAS CRASHED. PLEASE SEND ME YOUR EMAIL AND PRICES FOR THE PARTS. I WILL ALSO NEED A NEW SCSI CABLE FOR THE HARD DRIVE.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Probo Medical | We Buy, Repair, and Sell New and Used Ultrasound Equipment

We provide sales, service, rental, and repair for ultrasound systems, probes, and C-arms. We have the largest inventory in the world. Contact us about our special offers on certified refurbished equipment! Call: 317-759-9210 Email: sales@probomedical.com.

stats
advertisement

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)