Sujay Das
Auto technique does not work in OEC 9600
July 28, 2011 12:33
Hi,
We replaced the high voltage cable assembly in OEC 9600 recently to fix the collimator rotation problem but after replacing the cable assembly we are now facing a new problem.
Although the rotation problem is now resolved but the auto technique does not work properly. The KV shoots up to 120 kv and the image is completely bright. However the manual technique works fine.
We rechecked the cable connections and reseated them once again. We also put back the old cable assembly from outside ( without the HV connection at both the ends ) to rule out problem with the cable itself , but the problem still remains. Another observation was DS1 LED ( red ) glows on the Power Motor Relay Board ( located on the hinged door ).
Any help would be highly appreciated.
Thanks,
Sujay
to rate and post a comment
Juan de la Cruz Martinez
Re: Auto technique does not work in OEC 9600
July 28, 2011 10:54
Do you mean that the manual technique let you see a good image/ if that is the case the camera is not controlling the auto dose. If you do not see a good image in manual, the camera is not seen video , either the II or the camera has lost voltage.