please I am getting this error from Siemens Axiom Multix M machine.

Error 0010
Err Nr.0x0D System: Generator
16Bit-P.:0x0001

My email address is: adantene@gmail.com

Thank you

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: Axiom Multix Error 0010

September 02, 2013 11:59

which generator do you have?

Log inor Register

to rate and post a comment

(432)
(83)
(3)

juan pablo amado

re: Axiom Multix Error 0010

September 02, 2013 05:50

Error 010 Emergency stop triggered, release emergency switch. Safety circuit triggered. Check
longitudinal safety endswitches S41, S42 in 3D- TOP stand or S1, S2 in Multix Pro and
connection contactor K1.

Log inor Register

to rate and post a comment

(15)
(5)

MUHAMMED AHMED

re: re: Axiom Multix Error 0010

January 02, 2014 12:02

hi i have multix ease
its have error 0009
and its d510 board error
i need for repairing of board
i need tha scemetic drawing of d510 board
do u have plz send me........thankx

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: re: Axiom Multix Error 0010

January 02, 2014 06:15

I do not have schematics just block diagrams. I have seen this error before and normally it is a power supply low or defective. Here is the troubleshooting information.
Err009: D510 Error
Description:
· A D510 inverter error has occurred.


Possible Causes of Error:
· W520 cable defective or contact problems
· Power Supply defective
· D510 defective


Corrective Measures:
· Check low voltages (Wiring Diagram).
· Check the ribbon cable between D500 X10 and D510 X10 (W520); if needed, replace
· Replace the inverter


Err010: Power Supply error
Description:
· The generator has detected a power supply voltage error.


Possible Causes of Error:
· +15/-15 Volt low voltages too low or are missing.
· D500 defective.


Corrective Measures:
· Check low voltages (Wiring Diagram).
· D501 connections X7 and X8 are incorrectly connected. Causes a short-circuit of the 24 Volt.
· Measure the voltage directly at the test points on the D500 board. Alternatively, if no test points are available, the voltage can be measured at connector X1, D500. If the low voltages are present on the D500 board:
-> replace the D500

Log inor Register

to rate and post a comment

(15)
(5)

MUHAMMED AHMED

re: re: re: re: Axiom Multix Error 0010

January 03, 2014 09:30

hi
i need a schemetic diagram
and i was seen the d510 board
it line resistance 100ohm is defective
so i was change the resistance
but some smd component is blast how to check its value to change them
it is to need a schemetic diagram so if any one have diagram to send me plz
THANKS

1 person liked this
Log inor Register

to rate and post a comment